unmgvar » Sat Apr 25, 2009 6:43 pm
bobokiki,
your decision will have to be divided first according to those demands:
1. application needs. what are you doing what do you need to achieve?
2. cost to efficiency ratio- you can get faster, more sensitive, more reliable, ease of use, robustness, cheaper, but never all of them bundle together especially if you need to answer 1 as well.
3. System wide specs- check flow rate range, higher pressure limit, autosampler range of injection, type of injection, column oven range and column max size can add a pre-eluent or not. detector speed, amount of channels, system modularity, software capcities, GLP (if required), ease of use, calcualtions capacities
3. how much budget do you really have
4. tech/application support of your vendors. you do not want a white elephant in your lab.
when you answer that you will know if to go 1st for UHPLC or HPLC, and then 2nd if to go for a UV or PDA
as for your specific questions:
first of all remeber that there are still more column phases avalaible in analitical range and far more vendors providing possibilities. going UHPLC means less choice right now (it gets better every year)
not all vendors give the same specs for their UHPLC system, especially on 3 things. higher limit pressure of the system ; type of injection; speed of data collection of the detectors.
one interesting physical restriction is the 800 bar limit. that is where "plastics" become more "liquid" then solid. to be able to run at higher pressure means other materials (mostly ceramic- using buffers and working with "dry" system bring higher ware faster).
another interesting aspect of working above 800 bars is precision. precision of flow rate, precison of injection. i have seen an Accela not do 1 ml/min at 1000 bars, could do only 0.7 ml/min. you never know this unless you chekc it.
so even if you do go for a "super" resistant system, robustness is an issue needed to take into account. i have heard of UPLC customers that told that their Waters tech adviced them not to go above 800 bars because of the exponent ware it creates.
remember that sensitive is a double edge sword. sensitive results are achieved by more sensitive system with all of what it means.
Type of injection:
there are 2 injection types currently used that i know of in UHPLC systems. the difference affects performance, way of use and effective injectin range
split loops
pulled loops
split looped are the better ones, more precise better RSD. more expensive, require good engenering for low carry over (increases expenses), a lot less sample loss during injection, very fast injection time
pulled loops are less expensive, simpler engenering, if well used give by far the lowest carry over. but very high sample loss, need to be used as partial loops or full loop. very slow injection cycle (very important if you do very short runs like 1-2 minutes). RSD is less then in split loop but it does not mean that it won't be good enough for you, it depends on what you are doing
detector UV or PDA:
UV detectors on a single wavelenght give at least 7 times less noise(many times more even) then PDA- if you seek low detection limits this is an issue
UV detectors have a far better behavior of Refractive index value then PDA. this can mean smaller front peaks effects at zero dwell volumes, a lot less drift due to gradient methods.
PDA gives you 3D spectral information, althou in words it is short, but in information it means alot.
for both types of detectors you will need at least 40-60 Hz for a UHPLC system. do not bother with less if you go for 1-2 minutes runs and very narrow peaks
i agree with Bryan Evans on his suggestion and also advise you very strongly not to believe all of those sales people that will tell you that transfering from HPLC to UHPLC is easy. it requires validation. on the other hand if you start from nothing then it is not an issue you validate from scratch anyway.