Detector saturated...

Posted: Fri Apr 24, 2009 4:07 am
by brightsun
Hi everybody...

Give me some advice about this problem.

I am using GC/MS of Shimadzu (QP2100)

Now i'm anlysing PAHs in soil samples. When i run "Tuning", the software give me a notification " Detector saturated". (MS detector)

I'm trying to solve this problem by changing solvent cut time and diluting the extract, in the manual user's guide, there is another way: changing the detector voltage, but i can't understand this parameter.

Thanks for replying.

Have a nice day !

Posted: Fri Apr 24, 2009 7:00 am
by Sadanand
Hi,
your quality of soil samples may not be :idea: good,these may be more concentrated so you have to go on dilution untill you will get the normal reading; or if you are thinking to change the voltage parameter the help topic of a instrument can give us the complete idea.

Regards,
Sadanand Mallurwar

Posted: Mon Apr 27, 2009 2:56 pm
by Ron
If the detector saturated error occurs during tuning you probably have an air leak into the system.

Posted: Tue Apr 28, 2009 9:44 am
by brightsun
Ron is right, my system leaks.

Thanks Sadanand and Ron.