First of all, the flow rates used will depend on the column dimensions (and particle size, if pressure is an issue) and not on whether the particles themselves are totally or superficially porous. There are a number of on-line calculators that take the pain out of the calculations (you can even download one from our web site:
http://www.lcresources.com/sandbox/UHPLC-HPLC.xls ).
The other thing you have to match is chemistry, and that's not as straightforward. The place to start is with the USP column equivalents database:
http://www.usp.org/app/USPNF/columns.html . Scroll down to the "PQRI" database, since that includes more columns. My guess is that your best bet would be a 300A pore size column from Agilent.
Even with close surface chemistry, you will almost certainly have to adjust the mobile phase and gradient. I suspect your compounds (proteins ? ) will elute at a higher % organic.
Bottom line is that you are essentially developing a new method.