Calibration procedure, dedicated NP systems?

Posted: Fri May 08, 2020 2:01 pm
by DR
We have a pretty typical LC calibration procedure for all of our RP systems - water/0.3% Acetone for gradient formation and pump mixing tests, caffeine in water for wavelength accuracy, uracil, propyl- and methyl-parabens eluted on a C18 with ACN/water MP...

We have recently dedicated a system for NP only use (different seals) and we'd rather not have to convert it back to RP just for calibration.

So, does anyone have an alternative column and "go to" solvent for doing these same tests?

My initial thought is to substitute propanol for the water. This should allow the same (columnless) procedures to be used for the pump step/gradient formation tests as well as the detector wavelength accuracy.

If nobody sees a problem with that, what do you use for the precision and linearity? Parabens on a silica column with hexane/propanol or something?

Re: Calibration procedure, dedicated NP systems?

Posted: Mon May 11, 2020 1:13 am
by DR
Bump for Monday users...

Re: Calibration procedure, dedicated NP systems?

Posted: Mon May 11, 2020 2:04 am
by tom jupille
I had hoped someone else would have replied.

Those all sound reasonable to me. The only other thing I could suggest would be to check with the column manufacturer and see what they use as a QC test mix. That should be a reasonable choice for the precision and linearity tests.

Re: Calibration procedure, dedicated NP systems?

Posted: Thu May 14, 2020 12:00 pm
by DR
An excellent idea even if it's an example of thinking "in the box"! :lol: :roll: