vmu » Fri Apr 24, 2020 6:53 pm
Waters AccQ-Tag:
https://www.waters.com/webassets/cms/li ... 930494.pdf
https://www.waters.com/webassets/cms/su ... 052881.pdf
Waters AccQ-Tag Ultra:
https://www.waters.com/webassets/cms/su ... 001331.pdf
AccQ-Tag is an HPLC-UV or an HPLC-FLD variant. AccQ-Tag Ultra is a UPLC-UV variant (smaller particle size, higher pressure, higher speed of analysis). Both variants use essentially the same derivatization procedures and reagents. The derivatization is manual, but fairly simple. Use of internal standard is desirable to obtain accurate results. Derivatization kit for the HPLC variant is labeled as AccQ-Fluor because the reagents are tested for fluorescent impurities by the manufacturer. Eluent concentrates and columns are marketed by Waters. UPLC variant uses non-linear gradient in order to separate Lys and Tyr. Hence, Waters Acquity UPLC system is desirable. Agilent Infinity 1290 with ISET function in the software (emulation of other LC instruments) can be used instead. HPLC variant is relatively sensitive to the design of the eluent mixing unit of the instrument, which influences the stability of retention times of the early eluting amino acids.
Derivatizaton with OPA. HPLC-UV or HPLC-FLD. Solutions from Agilent. Variants with Hypersil AA-ODS column and with Zorbax Eclipse-AAA column. Derivatization is carried out in the autosampler of the HPLC instrument. FMOC reagent is added only when proline is needed.
https://www.agilent.com/cs/library/slid ... 0Steed.pdf
With Hypersil AA-ODS, it is sometimes difficult to obtain acceptable separation of Ala/Arg and His/Gly and to obtain acceptable peak shapes for Asp and Glu (addition of THF is the cure), but this is not a problem for your task. Phe and Ile are eluted close to each other, but the separation is usually acceptable.
If necessary, centrifugal filters can be used to remove the enzyme from your samples before the derivatization.