Yes, it is not uncommon for the background signal to increase after preventative maintenance, especially for lower masses. This can be due to a variety of factors such as residual cleaning agents, contamination from the maintenance procedure, or changes in the instrument settings. Here are some suggestions to help reduce the background signal:
Verify the instrument settings: Ensure that the instrument settings, such as the ion source temperature, ionization mode, and gas flows, are properly set for the specific application. Also, check the collision energy, declustering potential, and other MS parameters to ensure they are optimized for the desired analytes.
Check for contamination: Conduct a thorough check of the instrument and sample preparation process to identify potential sources of contamination. For example, check the sample vials, syringes, and other equipment for cleanliness, and inspect the mobile phase and solvents for impurities.
Run a blank: Before analyzing samples, run a blank to detect any potential contamination in the system. Use the same solvents and conditions as the actual samples to simulate the conditions of the analysis.
Increase the cleaning frequency: If the background signal persists, consider increasing the frequency of preventative maintenance, such as cleaning the ion source, replacing the ion source components, or changing the mobile phase and solvent bottles.
Contact the manufacturer or a service provider: If the above steps do not resolve the issue, contact the instrument manufacturer or a service provider for further assistance. They may be able to identify and resolve the problem.
Overall, it is important to conduct regular maintenance and quality control checks to ensure the instrument is operating at its optimal performance.