Is it safe to turn on the MS fillament if the source is cold

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
Hi all;

I am working with a Trace Ultra GC, coupled to a Finnigan Mass Spec. Yesterday, my boss and I changed the fillament in the ion source. Everything pumped down fine, but now the source refuses to heat up ( the read back is 22 C, Set point 200). The transfer line is at 250 C, so I suspect the issue is with the thermocouple, and not the heating element. The ion source itself is a combined EI/CI source, operating in EI mode.
Under these conditions, is there any risk of damage if I turn the fillament on ? I'd like to at least see if it is operating properly.

may be u did't fix wires properly

Thats a good suggestion.

I took the source out this morning, all of the wires were connected. However, one of the leads for the thermocouple was in electrical contact with the adjacent guide post. I corrected this and re-installed the source.

she's pumping down now, so we'll see what happens.

By the way, I found out it is safe to turn on the ion source. The heater is there primarily to keep larger analytes volitile ( thus preventing buildup). Its ok to look at the H2/N2/O2/H2O ions for a leak check. It would be unwise to inject something like methyl stearate (one of our standards).
3 posts Page 1 of 1

Who is online

In total there are 32 users online :: 0 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 32 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry