- Posts: 269
- Joined: Fri Jun 29, 2007 1:48 pm
To make a long story short, I am dissolving 5 grams of product (cream/lotion) into 15 mL of solvent. Not having any 15 mL volumetric flasks on hand I have been adding the placebo product, spiking a known amount via pipet (say 5 mL) and bringing up to volume with an additional 10 mL of solvent delivered by pipette.
All of my recoveries across the assay range (0.1% to 10%) are nearly always 25% to 30% low. I got concerned about this and began looking into the reasons behind consistently low recoveries across the assay range.
Long story short, I made up another sample and pipetted in 15 mL of solvent. I then transfered that into a 25 mL graduated cylinder and just as I began to suspect, I had about 20 mL of liquid.
Clearly I am suffering from volume displacment due to such a large sample size in combination with a small sample volume. There are two ways I can think of to fix this. The first is either to use a correction factor (but then I would need to determine the volume displacment over various weight ranges). The second is to use an internal standard.
Now for the questions after the long history: Would this be an acceptable case in which to use an internal standard? How ugly does this get if I want to say I would be using that same internal standard to quantitate unknown impurities?
I think this is a good case for the use of an internal standard. I am not so sure it is the best idea to quantitate unknowns based off an internal standard. Seems to me like I would need to know a good deal of information about my unknowns before I could really get away with this --- relative responce factor at the least.
Thoughts?
Thanks in advance!
Shaun