Dionex AS-40 Autosampler 5.0 mL vs 0.5 mL vials
I have been working on an older set-up using a Dionex AS-40 autosampler. The autosampler, itself is a work horse. I haven't had any trouble with it after I figured out the correct programming but have only had success injecting samples from 5.0 mL vials. If it try to inject from a 0.5 mL vials, the autosampler recognizes the 0.5 mL vial but results from standards are spotty at best. I do not get enough data to try and figure out what the issue may be.
Is there a programming change that I need to make to to adjust to the smaller volume injection? Or would there be a modification to the instrument that I would need to make, such as a tubing length change?
Thanks in advance for any input.
Hi lewright,
You may need to adjust the timing of the relay signal sent to the sampler by the IC. For 5 mL vials, the relay is usually closed at -2.3 min before the IC valve actuates. For 0.5 mL vials, the relay is usually closed at -1.0 min before the IC valve actuates.
Best regards,
John Guajardo
Product Manager
Thermo Fisher Scientific
Hi John,
Thank-you so much for the reply. I will give that a try.
In total there are 23 users online :: 0 registered, 0 hidden and 23 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 23 guests