My detector signal has suddenly dropped with a factor of 4. Could this be caused by the lack of make-up gas (e.g. nitrogen) for the detector? The retention-times havent changed.
My detector signal has suddenly dropped with a factor of 4. Could this be caused by the lack of make-up gas (e.g. nitrogen) for the detector? The retention-times havent changed.
Hello Martinz,

yes! Did You have a FID? If yes, You need make up gas because of the flame geometry AND the gas velocity from the jet into the flame.
A good make up gas for FID is N2 or Air. The flow is instrument depending and mostly between 25 and 45 ml/min.

He is for FIDs not a good choice.

rgds

Bernd
bm

Thanks! turns out some #$%& let a valve open. lost a N2 bottle in less than a day.
3 posts Page 1 of 1

Who is online

In total there are 11 users online :: 2 registered, 0 hidden and 9 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider], Semrush [Bot] and 9 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry