If you aren't set up for headspace, I ask whether your "sample" is to be the liquid squeezed from a wipe, or the moist wipe itself. Because most likely you'd be extracting into water for analysis.I am trying to figure out the best way to perform alcohol determination on alcohol pre-moistened wipes with respect to USP <611>. What would be the best way to extract the alcohol from the pads for quantification?
Anyone have experience with this?
Is room temperature headspace injection an option. Using a gas tight syringe on your autosampler should allow you to do room temperature analysis of the ethanol. You should be able to program the autosampler to sample higher up in the vial than normal.
You will need a suitable internal standard of course.
This is a low tech solution but might do the job
Kevin
Is headspace an option for you?
I have headspace as an option. In theory, would it be better to first start off with headspace of direct injection by extracting into water for analysis?If you aren't set up for headspace, I ask whether your "sample" is to be the liquid squeezed from a wipe, or the moist wipe itself. Because most likely you'd be extracting into water for analysis.I am trying to figure out the best way to perform alcohol determination on alcohol pre-moistened wipes with respect to USP <611>. What would be the best way to extract the alcohol from the pads for quantification?
Anyone have experience with this?
Be sure to use 0.5 microliter injection as the maximum, water expands a ton in the inlet.
Thanks for your reply.Interesting problem. Water in your sample - like if you extract the wipes - is tough on a GC system. The posters above are trying to indicate that with all of their comments about injection volume when the sample is substantially aqueous.
If you have a static headspace sampler, you might be able to just do it directly.
Known mass of wipe in the vial, seal, and analyze directly. This gets you out of the whole water mess. Another take on it would be to add a known mass of wipe and a known volume of water (keep the phase ratio in the vial known and constant). Since you're not injecting anything more than water vapor, you don't have to worry about the expansion of water in the hot inlet. If I were going that direction, I'd probably calibrate using the method of standard addition.
Typically the alcohol is expressed in v/v% or w/w% in solution which wetted the wipe.You have to be concerned about the mass of the wipe because that's the only way you can control your sample size.
How is the content of the ethanol in the wipes expressed in the specification? wt%, ppm? g/ft^2?
What if you simply extracted the wipe with acetonitrile? I'll bet the extraction would be pretty quantitative in 1 shot and it gets you out of the water problem for direct injection GC. I've seen where people use acetonitrile as an internal standard in EtOH analysis so there must be some chromatographic conditions to separate the two.Typically the alcohol is expressed in v/v% or w/w% in solution which wetted the wipe.You have to be concerned about the mass of the wipe because that's the only way you can control your sample size.
How is the content of the ethanol in the wipes expressed in the specification? wt%, ppm? g/ft^2?