Quantifying Low Concentration of active by HPLC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
we are working on a liquid dosage form in which active drug concentration is 2.5ug/ml. We have tried inecting a neat sample into the HPLC, DAD Detector but area count of impurities are very small so it is hard to detect and quantify impurities. we have tried several wave lengths to maximise the sensitivity of impurities but we can not. Is there any other way to maximise the sensitivity of impurities? pLEASE HELP..

FYI.
we can quantify impurites in the API with better sensitivity but we can not quantify impurities in the Finished product with 2.5ug/ml concentration.
Jey

If the dosage form is in a weak solvent such as water, you can inject a larger volume (I'm assuming reversed-phase chromatography). You will have to experiment to determine how high you can go; at some point overloading of your API will cause problems.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 18 users online :: 1 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry