- Posts: 16
- Joined: Fri Sep 05, 2008 6:34 pm
I have been running a split residual solvents method on GC to separate ethanol (10%), acetonitrile (410ppm) and DMSO (5000 ppm). However my boss wants us to switch to splitless because all our other lab sites do and we weren't getting great seperation between EtOH and ACN. So I started working on that and came up with a decent separation of all. However, when I started to run system suitability and some accuracy/precision runs DMSO failed with %RSD numbers of greater than 10% and also did not pass a linearity exp. I repreped all my samples and still the same happened so I do not think it's my sample prep.
My question is how to get DMSO to pass? Is it even possible? Here are my conditions for my new method. I'm still quite new to GC (only about 6 months experience) so I have a feeling I am overlooking something really simple here! Thanks in advance for the help!
GC: Hewlett Packard 5890 series 2 with FID and an autoinjector
Column: Restek Stabilwax, 0.53mm x 30m
Carrier Gas is Helium at 6.0 ml/min
Initial temp is 68C for 2 min
Temp ramp is 50C/min to a final temp of 210 for 1.5 min
The purge valve opens at 0.2 min
My sample is diluted 50 fold as well.
Let me know if you need anymore information! I also realized today we are still using the same split linear with a small bit of steel wool inside...I'm not sure if that could effect this. I also used to notice in the old method a lot of carryover of DMSO that would sometimes require me to make more injections or do blank water washes till it finally passed system suitability.