Same interval peaks

Posted: Mon Apr 13, 2009 6:33 pm
by DocHollywood
Hi,
I recently have been getting peaks eluting out at the same interval, even when i don't inject anything into my Plot U capillary column on my 5890 series 2 GC. There is basically an array of them which can small peak heights coming out at around every 40-50 seconds. Has anyone seen something like this before? If so, some direction would be useful before i start to take the GC apart.
Thanks
AD

Posted: Tue Apr 14, 2009 7:37 am
by Peter Apps
Please post an example - instructions are in a sticky at the top of the LC page.

Peter

regular peaks

Posted: Tue Apr 14, 2009 1:39 pm
by DocHollywood
Here is the image
Don't mind the fact the base line goes down initially that just occured on the run i saved because i had changed the pressure of hydrogen to the Mass Flow Controller. You can see the peaks come at regular intervals.
Image

Thank you for your time

Posted: Tue Apr 14, 2009 4:13 pm
by Peter Apps
This looks like electrical interference. Check around for any equipment that switches on and off at these intervals.

Peter

Posted: Tue Apr 14, 2009 4:15 pm
by larkl
If you change the temp gradient or the temperature (if it's isothermal), then you could tell if the peaks move. If they do, wouldn't it be some contamination? If they don't, then it's electronic.

Posted: Wed Apr 15, 2009 7:59 pm
by DocHollywood
Thanks for the quick response