how to seperate plasma interference from drug peak....

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

8 posts Page 1 of 1
I am confronting a problem of plasma front merging with my analyte peak.
Chromatographic conditions are:
mobile phase: MeOH:25mM phosphate buffer(38:62)
flow rate: 0.450mL/min
column: Genesis C18(100cm*4.6i.d.) 4µparticle size
oven: 40°c
Rt of my drug is 5.4....
With Regards
Shantaram

The usual suspects:
- change the aqueous/organic ratio
- change the temperature
- change the organic solvent from MeOH to ACN (or THF)
- change the pH
- change the column

There is no magic; you have to be systematic and persistent!
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

The usual suspects:
- change the aqueous/organic ratio
- change the temperature
- change the organic solvent from MeOH to ACN (or THF)
- change the pH
- change the column

There is no magic; you have to be systematic and persistent!
I had tried most of them except ACN and temperature. What about temperature should i increase to how much????
With Regards
Shantaram

What about temperature should i increase to how much????
I have no idea. :) Try something and see what happens.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Dear Shantaram,
You analyse intermediate lipohilic drug in serum. The front is more polar.
Therefore suggestion is:
1. Extraction plasma with methyl t. buthyl ether or methylene chloride or
ethyl acetate (manipulate with pH).
2. Column of 10 cm is much short for plasma, 25 cm or 15 cm is need.
3. I underrstand that peak symetry is good for plasma-free standard.

Dear Shantaram,
You analyse intermediate lipohilic drug in serum. The front is more polar.
Therefore suggestion is:
1. Extraction plasma with methyl t. buthyl ether or methylene chloride or
ethyl acetate (manipulate with pH).
2. Column of 10 cm is much short for plasma, 25 cm or 15 cm is need.
3. I underrstand that peak symetry is good for plasma-free standard.
I had tried with t. butyl methyl ether and ethyl acetate, but i will look for the larger column.
Thank you
With Regards
Shantaram

From your HPLC conditions - it seems that the drug elutes fairly early and maybe closer to the dead volume.

I would try a Phenyl column to see if it improves retention - at lower temperature. Next, try a polar embed phase and last resort will be HILIC.

From your HPLC conditions - it seems that the drug elutes fairly early and maybe closer to the dead volume.

I would try a Phenyl column to see if it improves retention - at lower temperature. Next, try a polar embed phase and last resort will be HILIC.
Thank you, i will try... :)
With Regards
Shantaram
8 posts Page 1 of 1

Who is online

In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry