METHOD DETECTION LIMIT
Posted: Fri Apr 10, 2009 10:07 am
by chemjeeorganic
Hi
I want to determine the MDL for PAH anlysis. I prepared 10 ppb standard. i am using a split injection (10:1). I wants to know whether can i concentrate this standard to 1ml and do the injection to do MDL.
Thanks
JN
Posted: Fri Apr 10, 2009 11:15 pm
by Don_Hilton
Personally, I would avoid concentration steps in making standards. If your PAH standard is too dilute, what about the idea of running splitless, which will give you a lower LOD?
Posted: Tue Apr 21, 2009 5:18 pm
by gpronger
Chemjeeorganic;
Could you give some more details on your purpose for the "MDL" and the whole analysis?
Thanks,
Greg
Posted: Wed Apr 22, 2009 2:12 pm
by randy
I would think that a 10ppb standard could easily be detected, and that dilution and not concentration would be the way to go.
MDL
Posted: Sat Jun 06, 2009 8:21 am
by chemjeeorganic
Hi Randy
Do you mean you can detect 10 ppb by SIR?. If so can you let me know how to achive it?
Thanks
NJ
Posted: Mon Jul 13, 2009 11:08 pm
by gpronger
Chemjeeorganic,
We perform a fair number of analyses by GCMS using SIM. We run in the splitless mode, and calibrate from 10PPB to 1000PPB injected concentrations. This gives a somewhat quadratic fit, but our sample concentrations tend to be fairly wide ranging, so we find it more efficient to run more calibration points (this is an 8 point calibration) to avoid dilutions. The instruments are an Agilent 5972 and an Agilent 5973. We have just begun evaluating our Thermo DSQ-2 for this application, and are able to calibrate to 0.5 ug/L (again in SIM).
Sincerely,
Greg Pronger
method detection limit
Posted: Tue Jul 14, 2009 3:51 pm
by chemjeeorganic
Hi Greg,
This was the information i was looking for. Already i have developed SIM method for SVOC analysis. But i was not sure whether to do spilt or splitless injection. Let me try i will let you know
Thanks
jn