I am currently validating a GC-MS for Trihalomethanes as per USEPA8260B but am facing a couple of issues...
I am able to run a good calibration but am having issues with a tendancy of some of the compounds withing the QC standard to read higher than expected...
I am using a 20ppb mix standard containing
Bromoform (mean result 26.2 SD 2.4)
Chloroform (mean result 23.8 SD 2.2)
Bromodichloromethane (mean result 23.4 SD 4.1)
Chlorodibromomethane (mean result 33.2 SD 6.5)
Thanks in advance as I would welcome any ideas or input in resolving this issue...

ninja,

Same vendor for the calibration standards and the QC standard? If not, what happens when you run the level 3 calibration standard as a sample? Same water source in both cases?

Best regards.

Try another mix, or prep one yourself if you have any neat compounds available. It seems as if one or the other mixes you are using are questionable.
Is your purge trap ok? For the Vocarb 3000 type traps, the brominated compounds fall out first.
3 posts Page 1 of 1

Who is online

In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: satya11, Semrush [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry