how to check linearity of knauer uv dector 40D....?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
how to check linearity of knauer uv for mplc and hplc liquid Chromotography system i am getting problem uv wavelength graph which is suddenly goes down in running the system
Check with Knauer and see if they have a specific procedure for that detector. In general, inject a series of standards of known concentration covering the range of interest. That will show you the linearity of the *system*. If you want to evaluate the detector only, the procedure is similar, but simply use a syringe to pull the solutions through the flow cell of the disconnected detector. You will have to do some experimentation to choose a standard that has an absorbance maximum at a wavelength you want, and the concentration of that standard to give you reasonable absorbance values.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry