How to accurately quantify Boron in ultra-pure water?
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
9 posts
Page 1 of 1
I am looking for methods to quantify trace Boron levels in ultra-pure water using ICP-MS. Can anyone point me in the correct direction?
-
- Posts: 247
- Joined: Sat Apr 28, 2007 3:24 pm
Should be relatively straightforward in ultra pure water matrix. Make some standards and go. Do you have experience with ICP? My table of polyatomic interferences shows Carbon-12 as an interference for boron, but I wouldn't think that would be much of a problem in ultra pure water. I have never actually tried it though, maybe someone has more practical experience?
-
- Posts: 95
- Joined: Mon Feb 12, 2007 10:05 pm
I run boron on an ICP OES and can usually get pretty consistent results down around 50 PPB using a Mod Lichte neb. Sometimes I'll see extremely high results but was told it was related to using borosilicate glassware. I only use plastic volumetric flasks etc to keep consistent standards.
It's pretty straight forward as OES at least.
It's pretty straight forward as OES at least.
-
- Posts: 51
- Joined: Tue Nov 04, 2008 8:17 pm
Boron has two isotopes with the masses 10 and 11. B11 is the most abundant and you should use that mass for quantification by ICP-MS. Therefore, Carbon-12 can't be an interference neither isobaric nor polyatomic.
You should be aware that Boron has a huge memory effect, you will have to rinse long time after your last standard before running the samples, especially at high concentration levels.
You should be aware that Boron has a huge memory effect, you will have to rinse long time after your last standard before running the samples, especially at high concentration levels.
-
- Posts: 37
- Joined: Fri Oct 05, 2007 4:07 am
Thank for your input guys. What is this memory effect of Boron? How will it affect my standards when i do calibration on ICP-MS? How will it affect my sample results?
Also how would i know if i have eliminated this memory effect before i run my samples?
Also how would i know if i have eliminated this memory effect before i run my samples?
-
- Posts: 247
- Joined: Sat Apr 28, 2007 3:24 pm
I realize that we would not expect C12 to interfere with B11; however, I am looking at a table of polyatomic interferences that lists C12 as an interference for B11. Next to C12 it says "spread". Not sure exactly what that means, but a ref is given (Clin. Chem. 1991, 37, 210-215) if you are interested.Boron has two isotopes with the masses 10 and 11. B11 is the most abundant and you should use that mass for quantification by ICP-MS. Therefore, Carbon-12 can't be an interference neither isobaric nor polyatomic.
-
- Posts: 37
- Joined: Fri Oct 05, 2007 4:07 am
C12 interference is not an issue to me because i am not expecting any contamination that is organic in nature.
-
- Posts: 51
- Joined: Tue Nov 04, 2008 8:17 pm
I had a look at the paper, I guess they meant that the instrument had poor resolution, this is not an issue with the instruments nowadays.I realize that we would not expect C12 to interfere with B11; however, I am looking at a table of polyatomic interferences that lists C12 as an interference for B11. Next to C12 it says "spread". Not sure exactly what that means, but a ref is given (Clin. Chem. 1991, 37, 210-215) if you are interested.
Concerning the memory effect, Boron tends to volatilize as boric acid from the sample solution layer that covers the inside surface of the spray chamber. If you rinse with an alkaline solution you can reduce that memory effect. If you write me at the email below I can send you a paper about Boron determination with ICP-MS where they use a special mixture to eliminate the memory effect.
-
- Posts: 37
- Joined: Fri Oct 05, 2007 4:07 am
Thanks Jerole. Please send it to joelyue@asiamail.com
9 posts
Page 1 of 1
Who is online
In total there are 50 users online :: 0 registered, 0 hidden and 50 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 50 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 50 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science