Sample analysis resolution vs standard resolution
Posted: Tue Apr 07, 2020 12:55 pm
				
				Hi there,
I'm trying to perform residual solvents testing of radiopharmaceuticals via GC-FID. When I inject my mixed standard containing Ethanol and Acetonitrile in water, there is some peak tailing being seen in the Ethanol peak.
When I perform injection of the sample, the peaks attributed to Ethanol and Acetonitrile are very poorly resolved (pretty much merging into another) and the Acetnoitrile peak is showing double humps. The sample is based on a NaCl matrix.
My current method parameters are based on a 30 x 0.53 x 3mm column, on N2 carrier gas (due to cost rises and availability of He), with a 250deg C FID and a 150deg C injection port. The column oven is isothermal at 40degC and the injection volume is 1uL.
Ive tried changing the liner,o-ring and septa to no avail.
Can anyone advise of any other steps I can take to try and mitigate the above sample issue? Im restricted by method stipulations in the Marketing Authorisation as per the below, but any guidance/recommendations would be appreciated!!!
Column: 30m x 0.53/0.32 x 1-3
Column stationary phase: Fused silica capillary column coated with cross-linked 6% of polycyanopropylphenylsiloxane and 94% polydimethylsilloxane.
Injector temperature: 150 - 250deg C
Injector volume: >=1ul
Oven pressure: 0.1 - 8bar
Flow rate: 1-15ml/min
Split ration: 5 (may be adjusted provided satisfactory detection)
FID temperature: 200-300degC
			I'm trying to perform residual solvents testing of radiopharmaceuticals via GC-FID. When I inject my mixed standard containing Ethanol and Acetonitrile in water, there is some peak tailing being seen in the Ethanol peak.
When I perform injection of the sample, the peaks attributed to Ethanol and Acetonitrile are very poorly resolved (pretty much merging into another) and the Acetnoitrile peak is showing double humps. The sample is based on a NaCl matrix.
My current method parameters are based on a 30 x 0.53 x 3mm column, on N2 carrier gas (due to cost rises and availability of He), with a 250deg C FID and a 150deg C injection port. The column oven is isothermal at 40degC and the injection volume is 1uL.
Ive tried changing the liner,o-ring and septa to no avail.
Can anyone advise of any other steps I can take to try and mitigate the above sample issue? Im restricted by method stipulations in the Marketing Authorisation as per the below, but any guidance/recommendations would be appreciated!!!
Column: 30m x 0.53/0.32 x 1-3
Column stationary phase: Fused silica capillary column coated with cross-linked 6% of polycyanopropylphenylsiloxane and 94% polydimethylsilloxane.
Injector temperature: 150 - 250deg C
Injector volume: >=1ul
Oven pressure: 0.1 - 8bar
Flow rate: 1-15ml/min
Split ration: 5 (may be adjusted provided satisfactory detection)
FID temperature: 200-300degC