In-situ isomerisation: calibration curve and LOD/LOQ
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
I have a compound that is isomerising in solution and the cis and trans isomers can be clearly separated by HPLC. However the trans isomer is about 98% while the cis isomer is about 2%. At low concentrations I am able to integrate the trans isomer easily but not the cis. From literature the equilibrium between the two isomers always stays the same providing the solvent system does not change. Is it acceptable for me to use only the larger trans isomer peak when generating a calibration curve and when determining my LOQ and LOD? If I can't my LOQ/LOD values are going to be very high!
-
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
If you are quantitating the total amount of the compound, and if, indeed, you can demonstrate that the equilibrium is 98/2, then yes, you should be able to use just the major isomer for LOD & LOQ estimation. LOQ (S/N = 10) effectively implies a CV of +/- 5%. LOD (S/N = 3) implies a CV of +/- 20%.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
If I had a method that was unable to detect 2% of my load, I would seriously consider a comprehensive revision of the method.
Best Regards
Best Regards
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
But if you have to determine LOD and LOQ for impurity profiling, and should reach 0.05%, then it's possible that 0.05% from 2% isn't visible while 0.05% from 98% still visible.If I had a method that was unable to detect 2% of my load, I would seriously consider a comprehensive revision of the method.
Best Regards
Ace
-
- Posts: 71
- Joined: Thu Jun 11, 2009 9:42 am
I'm not able to understand your thought to say this.
But if you have to determine LOD and LOQ for impurity profiling, and should reach 0.05%, then it's possible that 0.05% from 2% isn't visible while 0.05% from 98% still visible.
Ace
0.05% of 98% is always less than 2% of 100%.
If you're not able to see the 2% you'll never be able to see 0.05%.
Samuele Pedraglio
Developability Dept.
NiKem Research S.r.l.
Italy
Developability Dept.
NiKem Research S.r.l.
Italy
-
- Posts: 426
- Joined: Mon May 19, 2008 10:09 am
What I meant to say: if you have to look for peaks of 0.05% from your main peak, then it's possible that you can't see the 0.05% peaks from your main peak if the 2% peak is the main peak for your degradants you're searching for. So actually you are searching for peaks of 0.05% * 2% = 0.001% instead of 0.05% * 98%= 0.049%I'm not able to understand your thought to say this.
But if you have to determine LOD and LOQ for impurity profiling, and should reach 0.05%, then it's possible that 0.05% from 2% isn't visible while 0.05% from 98% still visible.
Ace
0.05% of 98% is always less than 2% of 100%.
If you're not able to see the 2% you'll never be able to see 0.05%.
I hope it's a bit clear now

Ace
6 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science