I have posted before of my trouble getting a stable RT with the silica-based amino column I have used and searching the resources has led me to some basic questions:
1. how common is it to use the amino columns with reverse phase solvents (acetonitrile/water) vs. normal phase solvents (EtOAc/hexane)?What are the advantages/disadvantages to each?
2. Should gradients be used or just isocratic run?
3. If used in an ion-exchange capacity, does the ion-ion interaction dwarf the H-bond (normal phase) interaction?
I am using this column to isolate what I think is a very polar low MW carboxylic acid (structure unknown). currently I'm using a gradient method that begins the run in 98:2 Acetonitrile/water and slowly introduces 20mM ammonium formate pH 2.8 in water to the flow. At about 20% B the compound elutes. My thinkning is that high organic content at the start of the run will allow retention of the compound and increasing aqueous content will elute it (if normal phase interaction is the mechanism of retention) and if it is an carboxylic acid then lowering the pH by introducing the pH 2.8 buffer will protonate the acid and cancel out any R-NH3+ -OOC-R ion-ion interactions. Compound retains well and separation from the matrix (soil extract) is good but the retention time is always shifting around sometimes by several minutes. Let me know if you see any obvious flaws here y'all.