I am having more peak area RSD problems. If I make successive injections of the same sample, my areas will often go up a little (2-5%) each time in a definite pattern.
I am trying to determine the concentration of an organophosphonate ester (bp 180-200 deg. C) in the 1mg/mL range in solutions of H2O, EtOH, DCM, and hexanes. The problem has occured with all solvents.
I have replaced the glass wool, tried different columns of the same type, cleaned my liner with acetone, DCM, etc. I have also tried decreasing my split ratio as well as cutting and replacing the ferrules on both columns.
If I remember correctly, I read something not long ago about reproducibility problems due to analyte collecting in the split vent and bleeding back out gradually. I'm thinking that either this is the problem or that there are active sites somewhere (liner or column) that are affecting elution amounts of my analyte.
One question that I have is -- what precisely does activity mean in this context? Just highly polar species or what? I ask because I think that injecting something guaranteed to not cause activity problems and observing the results would be a good troubleshooting step. Something along the lines of tetradecane maybe?
The area for 1mg/mL of my analyte also changes drastically when I use water compared to, say, DCM. Water gives areas around 30000, while DCM gives over 80000. We must deal with high concentrations and found our old split ratio (100:1) caused reproducibility problems.....though I suppose they could have been caused by the issue(s) described above.
I just wanted to bounce this off of you guys to see if anyone has had this experience and am super appreciative of any help you can give. Sorry for the novel and thank you.
Conditions/gear are as follows:
Shimadzu GC-17A chromatograph w/ JW Scientific DB-5 column (30m, .25 mm, .25 um). I am using a plunger-in-barrel 10uL syringe and making manual .5 uL injections.
Injector: 250 C; 120 kPa; helium carrier w/ total flow rate 31 mL/min; column flow rate 1.28 mL/min; avg. linear velocity 32.1 cm/sec; split ratio of 20:1. My FID is at 300 degrees w/ manufacturer-recommended flow rates.
Temp. Program: 100 C for 1 min, then a 40 deg. C/min. ramp up to 280 deg. C, then hold for 1 minute. My analyte elutes out just after 3.9 minutes.