Hi Guys,

I am doing Assay validation for byproduct of Raw material in my drug product. it is about 60% of active. Do i need to perform stress study of byproduct in my method validation? We are doing stress study af actives in drug product by different method. So my question is ICH or FDA asking for stress study of byproduct in method validation???

If you are stress testing the drug product, then you are stress testing whatever is in there (presumably, 60% byproduct and 40% API). Since the purpose of stress testing is to predict what you will find in stability testing, this is appropriate.

In other words, no, I see no point in stress testing the byproduct separately
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 595 users online :: 1 registered, 0 hidden and 594 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 594 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry