use of DCM

Posted: Thu Apr 02, 2009 10:39 am
by km06
Can we use Dichloromethane in supelcosil LC-18-DB column?

Posted: Thu Apr 02, 2009 1:09 pm
by Alex Buske
sure, why not? why would you?

Alex

Posted: Fri Apr 03, 2009 3:51 am
by km06
thanks alex,

one of our client wants to use this column with DCM and with a gradient program of,
Time (minutes) DCM, Methanol
0 10 90
20 10 90
55 50 50
60 50 50
60.1 10 90
70 10 90

Posted: Fri Apr 03, 2009 9:41 am
by lmh
Be careful if you have PEEK parts in your hplc system. I believe PEEK may have compatibility problems with DCM, but I might be wrong.

Posted: Fri Apr 03, 2009 12:21 pm
by Uwe Neue
I would ask the client to explain the method to me, and also I would ask him if he ever ran this method on more than one column.

Posted: Thu Apr 09, 2009 6:25 pm
by zokitano
Be careful if you have PEEK parts in your hplc system. I believe PEEK may have compatibility problems with DCM, but I might be wrong.
Imh has right. Mobile phases with % DCM higher than 5% can cause incompatibility issues with PEEK tubings or columns.

Posted: Fri Apr 10, 2009 10:22 am
by gcguy
If it is an Agilent system that you are using, the supply different piston seals for normal phase work.