Peak tailing and overloading

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

7 posts Page 1 of 1
I have a question considering overloading of a column. I have pretty concentrated sample (2 g/L) and I am making very small injections (10 nL). Do I overload column? My peaks tail. Would it be better if I was injecting bigger volumes with a way less concentrated sample? Let's say 5 uL injection of 0.004 g/L (the mass injected the same as in the above case). I am working with HILIC column (25cm, 4.6mm, 80A).
Thanks,
Anna
Anna Andrzejewska-Santiso

Anna,

From the injection voumes and sample concentrations you described, I don't think you either mass overload or volume overlaod the column. But I prefer using 5-uL injection of 0.004 g/L sample since most autosamplers or injectors will behave well. Also use mobile phase as the diluent. In some cases, peak tailing results from secondary interaction between the analyte and stationary phase, which can be oversome by using appropriate HILIC stationary phases and mobile phases.

What type of HILIC column are you using, neutral, ionic or awitterionic surface? And what is the nature of your analyte, basic, acidic or neutral?
Xiaodong Liu

Hilic column - TSKgel Amide-80 HILIC Column, so the stationary phase is mostly amide groups. It is based on polymer. I assume that amino group (-NH2) is cationic. As mobile phase I use mixture of (80/20 v/v) MeOH and buffer (pH=5 but in such a big amount of MeOH (80%) pH is shifted toward bigger values ~6-7). To be honest we use also CO2 in mobile phase (enhanced fluidity liquid chromatography). It acts as nonpolar solvent. My sample = cytidine (2 pKas, pKa1=3.3, pKa2=13.5). I assume that in this mobile phase (pH~6-7) it is positively charged (+1). Stationary phase is positively charged as well.
Sample is prepared in MeOH/Buffer solution (exactly the same as in mobile phase, except that mobile phase contains CO2).
BTW, not retained benzene tails as well. Is it a just a system characteristic? Thank you very much,
Anna
Anna Andrzejewska-Santiso

I'm not sure if a sub-critical CO2 system would be easy to model, especially as a HILIC system, and I would not be surprised if the CO2 interacted with the water. I would guess that the whole CO2 system is the cause of the tailing, unless you previously had good peak shape.

You could probably try the column in a standard HPLC system using a suitable HILIC mobile phase to ensure that the column provides good peak shape for your compounds, and then move the analysis to your system.

You could also contact Tosoh Bioscience support to ascertain if they have any useful comments.

Bruce Hamilton.

Thank you for hints.
We used reversed phase column (C18) for that system (of course there was different percentage of MeOH versus buffer) with CO2 and no such a tailing was observed. I figured out that this is the characteristic of the column. Packing heterogeneity probably causes tails.
Thanks,
Anna
Anna Andrzejewska-Santiso

Lots of things could cause tailing, and I assume the column works OK in normal HILIC HPLC mode, which may exclude packing issues - unless subjected to out-of-specification environments.

I'm not sure that a standard RP C18 column would be a good model for HILIC in your system, but it certainly does exclude many other possible systemic causes of tailing.

Bruce Hamilton

It won't be such a bad idea to make sure the column is working properly. You can do so by running vendor suggested column performance test on a "known" good LC system. It is possible somehow the column was damaged. Tailing of benzene suggests system and/or column problems. If you are sure the system works ok, it is more likely the column is faulty.
Xiaodong Liu
7 posts Page 1 of 1

Who is online

In total there are 59 users online :: 2 registered, 0 hidden and 57 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Semrush [Bot] and 57 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry