Autosampler temperature
Posted: Tue Mar 24, 2020 5:05 pm
by urosuros
Hi I have a method with a sample temperature 4 degrees +- 1 degree. I will be doing that method on waters alliance sistem and waters arc sistem. Can this alarm band +-1 degree be a problem.
Thank you
Re: Autosampler temperature
Posted: Tue Mar 24, 2020 6:49 pm
by DR
Yes.
4°C is as low as the sample compartments will go under good conditions and ±1°C is pretty tight.
Anticipate alarms. That said, I think you can tell the Alliance to ignore alarms via Empower...
If having sample temp. controlled that tightly is not really required, loosen it up a little and make your life easier.
Re: Autosampler temperature
Posted: Wed Mar 25, 2020 10:58 am
by urosuros
Yes I think that alliance sistem only shows warnings but doesn’t stop run sample but I don’t know about waters ARC sistem if it stops injecting sample when temperature is over alarm band