Rapid resolution HPLC pump
Posted: Fri Mar 27, 2009 7:42 pm
by PigmentQueen
Can anyone explain to me why an Agilent 1200 RRLC system sustains higher pressure than an 1100?
Posted: Fri Mar 27, 2009 9:14 pm
by Kostas Petritis
I do not know what are the exact modifications but for sure they must have upgraded the injection valve and made general improvements in their autosampler...
Posted: Fri Mar 27, 2009 11:48 pm
by Bruce Hamilton
Could you please clarify?.
Do you mean that for the same mobile phase flow, the system pressure is higher - if so that's probably due to using narrower bore components intended to reduce the 1200's internal volume.
If you meant, why can 1200 safely pump to a higher pressure, as noted by Kostas, that's due to the redesign necessary to compete with the Waters UPLC and similar high pressure instruments. Basically lower internal volumes, more accurate low flow settings, higher pumping pressure etc.
Bruce Hamilton
Posted: Sat Mar 28, 2009 12:57 am
by Kostas Petritis
I wouldn't consider the Agilent 1200 (max pressure: 600 bar) an equal to the Waters UPLC (max pressure 1000 bar)...