Retention time calculations for a different column.
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
2 posts
Page 1 of 1
I am trying to validate use of 150-x4.6 mm RPHPLC column instead of 125x4 mm column (same stationary phase), with gradient elution. If this was an isocratic separation, the correlation coefficient would be [(150/125)X(4.6/4)^2 = 1.587 calculated based on linear velocity. But I have a gradient.... and it seams to me that the dwell time should be involved as well. Are there calculators available on a web? Is there a way to use a DryLab for that? Thanks...
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
Neglecting the delay volume, just scale the flow rate with the square of the i.d.. Then scale the gradient volume, including the subsections of the gradient, with the ratio of the column volumes.
2 posts
Page 1 of 1
Who is online
In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 24 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 24 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science