Surrogate response factors
Posted: Wed Mar 04, 2020 1:52 pm
				
				I am running drinking water VOC's on an Agilent 7890/5977 GC/MS and an ATOMX purge and trap autosampler. For the last several months I have been having an issue with my surrogates. When I run my curve the responses of my surrogates increases as the standard concentration increases. My curve samples pass just fine but it puts the "average" of my surrogate responses high enough that the surrogates in my samples and blanks (which don't have a whole lot of compounds in them) are reading to low and not passing my QC (they need to be between 80 and 120% recovery and mine are down around 75%). The P&T people are blaming the GC/MS people and the GC/MS people are telling me it is a P&T issue. I don't know what. I've tried cleaning the source, changing the liner, cleaning the mount. I replaced all of the tubing in the P&T to the GC twice but I can't find anything that fixes the problem