Page 1 of 1

5890 Fault

Posted: Mon Mar 23, 2009 5:24 pm
by pueblito
Hello everybody.

my 5890A´s display suddenly showed:

FAULT:
ADC OFFSET.

I read the manual and I found:
"Indicates a problem with the thermal ADC offset reading. Check R168 and R172 in the main board".

but before the problem another one had been happening: an unusual base line:

Image
the detector was turned off.

Are these problems related each other?
Does the problem have solution without buying another mainboard? I mean, replacing components instead.
any suggestions will be appreciated.
many thanks


5890 fault

Posted: Wed Apr 01, 2009 8:19 pm
by oncoming storm
Pueblito,

If the problem persists, you need a new mainboard. The baseline you are seeing could be an indication of electronic noise created by your system. Although I have seen this kind of baseline with a detector off and so long as it doesn't affect my chromatography, I don't worry about it.

Hope this helps,
oncoming storm

Posted: Thu Apr 09, 2009 12:15 am
by pueblito
oncoming storm.
thank you for your opinion.