Here are the chromatogram:
Compond (increasing RT): Phenol, 4-Nitrochlorobenzene, Toluene, Naphthalene
Upper: Agilent Eclipse XDB-C18 4.6x150mm, 5um column (65:35 MeOH:H2O, flow:0.4ml/min, inj vol: 5ul)
Lower: Agilent Eclipse XDB-C18 2.1x50mm, 1.8um column (65:35 MeOH:H2O, flow: 0.21ml/min, inj vol: 1ul)

The chromatogram are silimilar to the LC column performance report shipped with the column.
When I add a bit more compounds (especially early eluting ones). I got the following 2 chromatograms:
Compound: Acetone, Phenol, Acteophenone, 3,4-Dimethylphenol, 3,5-Dimethylphenol, Propiophenone, 1,3-Dimethoxybenzene, 4-Nitrochlrobenzene, Benzophenone, Toluene, alpha-chloro-p-xylene, Naphthalene, Xylene
Upper: 70:30 MeOH:H2O, flow: 1ml/min, inj vol: 5ul
Lower: 70:30 MeOH:H2O, flow: 0.21ml/min, inj vol: 1ul
1. All tubings are reduced to small (0.12mm id)
2. Small volume flow cell (5ul) is used.
3. Injection solvet: 50:50 MeOH:H2O
4. Hardware: Agilent 1200 Quat Pump, DAD 40Hz, Standard Autosampler
Is the instrument not suitable for fast-LC column? (still dead volume somewhere?). Note also that there are small tailing on every peaks in the last chromatogram.