What is the effect of “high Temperatureâ€
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
At low pH, bonded phase is cleaved by acid catalyzed hydrolysis of the silyl ether linkage. These columns are arguably more stable than would be expected based on the stability of that bond, and one explanation is that the hydrophobicity of the bonded phase tends to keep it in place long enough for the bond to re-form (most of the time!). C8 is less hydrophobic than C18, and so less stable in a high temperature / low pH environment.one of my question still remains...why the difference in temperature limit between Zorbax SBC8 (which has temp limit of 80oC) and Zorbax SB C18 (temp limit of 90oC)
Contact your local Agilent agent and ask for their literature on "sterically protected" bonded phases. All will be explained there.also Temp tolerance of Zorbax sable bond (SB) is higher than other Zorbax columns such as Zorbax extented ...why so??