in agilent 7890 it gives me that the septum purge flow is in negative and could not reach the setpoint which is 3ml/min. what do u think the resone is? the front inlet pressure is 12 psi and the injector temp is 250, the column is hp-5ms, 30m, 25um, 0.25um film and the flow is He 1 ml/min
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
What is your split flow?
What is your carrier gas pressure at the cylinder regulator?
What is your carrier gas pressure at the cylinder regulator?
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Are you sure that you have the column diameter correct - I would expect 250 um, not 25 um. Then, depending on the column temperature the inlet pressure is too high for the flow rate - check that you have selected the right carrier gas.
Peter
Peter
Peter Apps
3 posts
Page 1 of 1
Who is online
In total there are 45 users online :: 0 registered, 0 hidden and 45 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 45 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 45 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science