Improving Mass Spec linearity

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hello,

Are there any tricks to improve calibration curve linearity? I typically run standards from 10 to 10,000 ng/ml and usually have to drop my two highest points do to possible over-saturation even though my ULOQ only has counts around 100,000. I use a triple quad for sample analysis and see this phenomenon across different molecular compounds.

Thanks for the suggestions.

A couple of questions:

Do you see this in neat solution as well as in extract?
Have you tried a different ionization mode?
Which triple are you using? When running the API5000 I typically see saturation at 500ng/mL, which is why I prefer a 4000 or even a 3000.

Even still, 100,000 counts at 10,000ng/mL is very low. Are these compounds typically hard to ionize (how does the tuning go?)

Finally, have you been able to replicate this phenomenon on a different instrument?
2 posts Page 1 of 1

Who is online

In total there are 98 users online :: 0 registered, 0 hidden and 98 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 98 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry