plate number counts discrepancy on normal phase column
Posted: Tue Mar 17, 2009 10:03 am
i have the same normal phase columns that are obtained from two vendors.
i believe they are very similiar for a simple 2-compound test mixture elute the same pattern from both columns under same isocratic conditions. only on column 2 they come out a little earlier.
the issue is by the calculations of the same plate counts, the sharper peak 1 has higher counts on column2 than on column 1, but the later eluting peak2 that is relatively broader is the other way around, the plate counts for it is higher on colum1 than column2.
i am wondering what this means for the column efficiency, are they supposed to follow the same trend? i can see the peak 1 is narrower on column2, while the peak 2 is less so. two peaks is about a minute apart. so peak broadening shall not affect that much.
does this mean the packing quality is different? i actually like the colum2 although peaks elute a littler earlier but the resolution is actually better. since this is normal phase column, does this mean the silanol activity is less on column2?