Thanks Consumer Products Guy.
More problems discovered today. The active inlet valve (AIV) gold seal and cap disappeared!!!

Some previous groupmates forgot to replace a new one in it (5years ago).
Hence I borrowed a pair from other research group. And used Channel B/C/D only. The baseline variation reduced from 100mAU to become 10mAU now. (Is it ok if I keep this MCGV but not using the leaking Channel A?)
Then I disconnected the tubing from pump to flowcell. Set it to higher flow rate (3ml/min). I saw the flow is not a continous stream, but as a stream for first stroke and droplets for the second stroke.
I suspected there is blockage in the inlet/outlet check valve. And 1050 AIV is "all-in-one" design! There is no cartridge inside the AIV as 1100/1200. Of course, I cannot drop the whole AIV into solvent and sonicate it.
I phoned Sevice Engineer and he suggested to run the system with 3N nitric acid, hoping that dirt will be washed off. I didn't try yet becasue it is already day-end today.