Ph.Eur. Prednisolone - Related Substances
Posted: Sun Jan 26, 2020 12:19 pm
by Philip01
Hi all,
We're following the Ph.Eur. monograph for Prednisolone but are having trouble in meeting the peak-to-valley ratio between Prednisolone and Impurity A (Hydrocortisone) for the Related Substances test.
The column detailed within the monograph is:
Length: 15cm
Internal Diameter: 4.6mm
Stationary Phase: C18 (End-Capped) 3μm
Instrument Parameters:
Mobile Phase A: Water
Mobile Phase B: ACN, Methanol (50:50 v/v)
Column Temp: 40 degrees
Flow Rate: 1ml/min
Injection volume: 10μl
Gradient table:
T0 - T14 (MPA 60% and MPB 40%)
T14 - T20 (MPA 60 -> 20% and MPB 40 -> 80%)
T20 - T25 (MPA 20% and MPB 80%)
The reference solution is a CRS vial containing these impurities made up in solvent mixture, Acetonitrile, Water (40:60). The requirements of the monograph is to have a peak to valley ratio of not less than 3.0 which we've been unable to meet. We're running on a Waters Alliance HPLC with a Luna C18 column (new) and are only able to get a p/v ratio of ~1.8 at best. We've also tried changing some of the instrument parameters that are allowed such as increasing the column temperature which did help a little.
Any advice would be greatly appreciated, thanks.
Re: Ph.Eur. Prednisolone - Related Substances
Posted: Sun Jan 26, 2020 3:12 pm
by vmu
Some useful info can be found here:
https://extranet.edqm.eu/4DLink1/4DCGI/ ... w/mono/353
Column: Venusil AQ C18.
Posted: Sun Jan 26, 2020 3:34 pm
by Philip01
Thanks, I have been looking at the knowledge base and came across the column they referenced. However would a particular brand make such a difference?
It's something we haven't tried yet though. Another option we're considering is if a H-Class UPLC might help due to a smaller dwell volume.
Posted: Mon Jan 27, 2020 5:15 am
by vmu
I don't think the dwell volume is the cause since the problem peak pair is eluted in the first isocratic part
of the run. However, insufficient equilibration time before the next run may cause problems with the separation. Check this.
Also check your data collection rate. The rate might be too low.
Posted: Mon Jan 27, 2020 5:23 am
by vmu
The column in the monograph info is an "AQ" column. This can change selectivity enough in comparison with Luna C18.
You can also try to decrease the injection volume if the detection sensitivity is sufficient.