Problem of suppressor for IC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
I cannot adjust the flow rate ratio between suppresor to conductivity detector and suppressor to waste. My suppresor is 828 IC Dual Suppresor (Metrohm). In the manual, the flow rate from suppressor to detector is supposed to be 40% from column flow rate. That means the flow rate from suppressor to waste is supposed to be 60% of column flow rate. I cannot reduce the flow rate from suppressor to column less than 70% of column flow rate. When I tried to do it, the suppressor showed not-ready status (red light). I need help from experts or any one who encounters the same situation. Thank you very much if you don't mind giving advice.

Dear Aopeau

Please make sure that you connected the back-pressure capillary (6.1803.090 PTFE capillary 0.25 mm i.d. / 15m ) after the detector. Only if this cappillary is connected there will be enough backpressure to get the correct flow ratio.
Dr. Markus Laeubli
Manager Marketing Support IC
(retired)
Metrohm AG
9101 Herisau
Switzerland
2 posts Page 1 of 1

Who is online

In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry