How select column to analyse O2 and N2 in CF4.thank.
i want make friends with whom interests chromatography.
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
You have to determine how much oxygen and nitrogen are in your CF4, or at least choose the lower limit you wish to measure of these trace (?) gases.
It is not as straightforward or as easy as many might think.
It would take too long to go through all the details and the choices involved and the problems involved at different levels.
Talk to Supelco or Alltech or Restek. Get professional help.
best wishes,
Rodney George
consultant
chromatographer1 at aol dot com
It is not as straightforward or as easy as many might think.
It would take too long to go through all the details and the choices involved and the problems involved at different levels.
Talk to Supelco or Alltech or Restek. Get professional help.
best wishes,
Rodney George
consultant
chromatographer1 at aol dot com
-
- Posts: 82
- Joined: Sat Sep 09, 2006 1:44 am
What kind of detector are you using? what level O2 N2 are you measuring?
It would normally be possible using an injection valve with a pre-column and a heartcut valve to get rid of CF4 and a molecular sieve to allow O2 N2 separation.
I would be able to measure from low ppb up to ppm using micro PED detector.
Let me know if you would need more support and I can push to find the right pre-column. Maybe porous polymer but I'm not sure.
It would normally be possible using an injection valve with a pre-column and a heartcut valve to get rid of CF4 and a molecular sieve to allow O2 N2 separation.
I would be able to measure from low ppb up to ppm using micro PED detector.
Let me know if you would need more support and I can push to find the right pre-column. Maybe porous polymer but I'm not sure.
-
- Posts: 82
- Joined: Sat Sep 09, 2006 1:44 am
What kind of detector are you using? what level O2 N2 are you measuring?
It would normally be possible using an injection valve with a pre-column and a heartcut valve to get rid of CF4 and a molecular sieve to allow O2 N2 separation.
I would be able to measure from low ppb up to ppm using micro PED detector.
Let me know if you would need more support and I can push to find the right pre-column. Maybe porous polymer but I'm not sure.
It would normally be possible using an injection valve with a pre-column and a heartcut valve to get rid of CF4 and a molecular sieve to allow O2 N2 separation.
I would be able to measure from low ppb up to ppm using micro PED detector.
Let me know if you would need more support and I can push to find the right pre-column. Maybe porous polymer but I'm not sure.
4 posts
Page 1 of 1
Who is online
In total there are 39 users online :: 0 registered, 0 hidden and 39 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 39 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 39 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science