- Posts: 1
- Joined: Wed Mar 11, 2009 5:51 pm
Story is :
1, I extracted lipids with petrolium either and saponified to get residue of the unsaponifiable matter for e.g a total of 0.09 grams was separated and this was diluted in 2ml dichloromethane.( this is my sample to run in GC)
2. I have cholesterol with 1200ppm, Sitisteriol with 1500ppm for the internal standard to prepare and I have a total volume of 1.5ml vial to fill for the GC. what I did was I prepared 6 factors with dilution factor of , 5x, 7.5x, 15x, 25x, 75x, 150X,
that is is 300 +1200=1500ul that is 5x dilution factor
200 + 1300= 1500ul
100 + 1400= 1500ul
60 + 1440= 1500ul
20 + 1480= 1500 ul
10 + 1490= 1500ul .....150x dilution factor... in the GC soft ware i can see that some of the peaks are in the range but some of them are bigger peak that is big area so I still need to dilute my samples but now at the end i dono how to calculate back my unknown samples concentration , would u plz suggest anything or give me simple example to flow...
Regards,
Z.