Agilent 7890 - Inlet Pressure Shutdown
Posted: Sat Jan 18, 2020 8:17 pm
Hi all,
I've been working with a USP Pharmacopoeia method for the determination of impurities but I've run into a few problems with the inlet pressure shutting down.
The method states a linear velocity of 25cm/s (at 50°C) with the carrier gas being Helium which is what we're using. The column details are, 30m, 0.32mm and a film thickness of 0.5μm.
The inlet conditions were setup as follows;
Temperature: 150°C
Column Flow: 1.4ml/s
Mode: Splitless
Septum Purge: Off
Purge Flow: Off
Purge Time: Off
With these conditions the inlet pressure is preset at ~7psi but the pressure would shutdown as the inlet pressure was unable to rise to this preset value. After some research on this forum and understanding the theory of GC inlet systems I have set the following conditions;
Temperature: 150°C
Column Flow: 1.4ml/s
Mode: Splitless
Septum Purge: 3ml/min
Purge Flow: 30ml/min
Purge Time: 0.5
The pressure is easily maintained with these conditions but I'm wondering if this is the correct approach to take? On a side note after performing a SST injection it was noted that the early eluting peaks had a symmetry factor of <1.5 but a peak just before the solvent peak showed shouldering and a poor symmetry factor of ~3.5. I'm not sure if my Purge Flow / Purge Time is a factor in this?
Thanks.
I've been working with a USP Pharmacopoeia method for the determination of impurities but I've run into a few problems with the inlet pressure shutting down.
The method states a linear velocity of 25cm/s (at 50°C) with the carrier gas being Helium which is what we're using. The column details are, 30m, 0.32mm and a film thickness of 0.5μm.
The inlet conditions were setup as follows;
Temperature: 150°C
Column Flow: 1.4ml/s
Mode: Splitless
Septum Purge: Off
Purge Flow: Off
Purge Time: Off
With these conditions the inlet pressure is preset at ~7psi but the pressure would shutdown as the inlet pressure was unable to rise to this preset value. After some research on this forum and understanding the theory of GC inlet systems I have set the following conditions;
Temperature: 150°C
Column Flow: 1.4ml/s
Mode: Splitless
Septum Purge: 3ml/min
Purge Flow: 30ml/min
Purge Time: 0.5
The pressure is easily maintained with these conditions but I'm wondering if this is the correct approach to take? On a side note after performing a SST injection it was noted that the early eluting peaks had a symmetry factor of <1.5 but a peak just before the solvent peak showed shouldering and a poor symmetry factor of ~3.5. I'm not sure if my Purge Flow / Purge Time is a factor in this?
Thanks.