Hi everybody...

Give me some advice about this problem.

I am using GC/MS of Shimadzu (QP2100)

Now i'm anlysing PAHs in soil samples. When i run "Tuning", the software give me a notification " Detector saturated". (MS detector)

I'm trying to solve this problem by changing solvent cut time and diluting the extract, in the manual user's guide, there is another way: changing the detector voltage, but i can't understand this parameter.

Thanks for replying.

Have a nice day !

Hi,
your quality of soil samples may not be :idea: good,these may be more concentrated so you have to go on dilution untill you will get the normal reading; or if you are thinking to change the voltage parameter the help topic of a instrument can give us the complete idea.

Regards,
Sadanand Mallurwar

If the detector saturated error occurs during tuning you probably have an air leak into the system.

Ron is right, my system leaks.

Thanks Sadanand and Ron.
4 posts Page 1 of 1

Who is online

In total there are 119 users online :: 1 registered, 0 hidden and 118 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Semrush [Bot] and 118 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry