Dear Ram,
Because you're planning to analyze volatile non-polar analytes with molecular mass smaller than 300 Da, yes, it is recommended to use PDMS fiber with 100um thickness. PDMS coating is very suitable for wide range of non-polar volatile and semivolatile analytes. The distribution process of the analytes into the liquid PDMS coating is predominantly governed by absorption rather than adsorption. Generally, 100um PDMS coating is used when one plans to do analysis of volatile substances with molecular mass smaller than 300 Da and with distribution (partition) constants, K, smaller than 10000.
Also, when extracting small volatile molecules like your terpenes and sesquiterpenes, the 100um PDMS coating will give you higher enrichment factors and hence, lower detection limits. You could extract your analytes with 30um PDMS also, but you'll never get the enrichment factors for 100um PDMS, but you'll definitely cut the equilibration time to smaller values (times), required for establishment of the partition equilibrium between the analyte concentrations in headspace and the PDMS coating.
So, as you see every decision comes with some "trade-off". Picking the right fiber coating thickness is determined by your goals, properties of the sample and the sample matrix, properties of your analytes, the concentration of your analytes in the sample to be analyzed, time, cost of the analysis...
As you already know, the other SPME method parameters must be optimized after the picking of the SPME fiber coating (and thickness) like: vial size, sample and headspace volume size, temperature of extraction, pH of the sample, the addition of salt (salting out effect), equilibration time, etc. All these parameters must be adjusted also, if you're doing liquid-phase microextraction (or single-drop microextraction), as I understand from your previous post, that you're actually familiar with it.
Some references, which I hope will be helpful for you:
http://www.sigmaaldrich.com/Graphics/Su ... 0/4547.pdf
http://www.sigmaaldrich.com/etc/mediali ... /10942.pdf
Good luck with your SPME analysis