Posted: Thu Dec 26, 2019 11:03 pm
I'm running a PTV inlet on a 6890 with an FID, and recently had to change from the septum head to the septumless head (due to thread stripping on the septum head, hooray). At the same time I took the opportunity to clean out the vent line (to my knowledge had never been done!) and do all the usual cleaning and maintenance. Now, when I inject, I get a tailing solvent peak, the flat baseline for a few minutes... then a huge jump in baseline that slowly tails away over the next 10-20 minutes, the shape of a shark fin moving to the right. The timing of the jump is ~ the time the gas saver comes on + transfer time for the column (and moving the gas saver time moves the jump accordingly). So, it appears to be solvent that isn't getting flushed out of the inlet during the purge.
In the past, apparently this has shown up (without the septumless head involved) and was cured with increased purge flow. Increasing purge flow now isn't fixing this, though. Increasing purge flow/increasing purge duration doesn't change the occurrence of the baseline jump, just moves it out in time. The peaks themselves look ok.
New liner. System is leak tight.
Injecting in toluene, 60m vf17 column with 2m of retention gap. Helium at 1.1 ml/min. PTV starts at 110, instant-ramps to 320. 100 ml/min purge flow comes on at 2 minutes. Gas saver to 20 ml/min comes on 5 minutes after that. GC oven starts at 120. Lowering or raising the initial oven/PTV temps to be more noticeably above/below the boiling point of toluene hasn't helped.
Am I looking at an obstruction somewhere in the vent system? Might there be an issue at the EPC? The liner is new, but the inlet itself has been through a LOT - not too clean and shiny in there, despite solvent rinsing it. Can oxidization/buildup inside the inlet body hold onto solvent?
In the past, apparently this has shown up (without the septumless head involved) and was cured with increased purge flow. Increasing purge flow now isn't fixing this, though. Increasing purge flow/increasing purge duration doesn't change the occurrence of the baseline jump, just moves it out in time. The peaks themselves look ok.
New liner. System is leak tight.
Injecting in toluene, 60m vf17 column with 2m of retention gap. Helium at 1.1 ml/min. PTV starts at 110, instant-ramps to 320. 100 ml/min purge flow comes on at 2 minutes. Gas saver to 20 ml/min comes on 5 minutes after that. GC oven starts at 120. Lowering or raising the initial oven/PTV temps to be more noticeably above/below the boiling point of toluene hasn't helped.
Am I looking at an obstruction somewhere in the vent system? Might there be an issue at the EPC? The liner is new, but the inlet itself has been through a LOT - not too clean and shiny in there, despite solvent rinsing it. Can oxidization/buildup inside the inlet body hold onto solvent?