Waters Acquity H class
Posted: Thu Dec 19, 2019 6:02 pm
by Waters0709
I need help with cleaning. Is there anyway to clean the INSIDE OF THE SYRINGE WITHOUT taking it apart? I understand there is wash solvent and purge solvent. But h class only cleans outside. How do you clean inside of the needle? Without drawing sample and generating data.
Re: Waters Acquity H class
Posted: Fri Dec 20, 2019 4:26 pm
by Prep1
Prime the syringe from the console.
Re: Waters Acquity H class
Posted: Thu Jan 23, 2020 4:27 pm
by BrianUtri
UPLC HClass has a flow trought design, in that way, to clean the inside of the needle you just need to pump the solvent you want and you are good to go.
To clean the seringe you go to UPLC Console>Sample Manager FTN>Control>Prime.