Peak Shape problem
Posted: Sun Mar 01, 2009 2:54 pm
by Santosh Gandhi
Iam developing a hplc method for monitoring reaction . The starting material (7-Phenyl acetamido-3-Hydroxy-3-Cephem-4-Carboxylic acid diphenylmethyl ester i was told that keto -enol tautomerism is possible) gives better peak shape only at pH below 7 ( at pH above 7 the peak tails badly [it drags] , but the intermediates formed degrades at pH below 7. how to go about.
Thank you
Santosh Gandhi
Posted: Sun Mar 01, 2009 3:56 pm
by JGK
How exactly are you monitoring the reaction?
Are you trying to quantify all items Including the intermediates?
or
Just measuring decease in starting material/increase in product?
If it is the last option does the degradation of the intermediates affect chromatography?
Also how long does it take the Keto - Enol tautomerism to reach equilibrium?
(I am asking as I once had a an assay where the analyte took 2 hours to reach a tautomeric equilibrium. Attempting analysis before this point gave very poor calibrations and irregular peak responses).