i am in the middle of war of words at the moment regarding one of our HPLC instruments system suitabilty method.
The method in question involves the analysis of marine oil via GPC. Their is a system suitabilty standard which caontains a mixture of TG-DHA, DG-DHA and MG-DHA (which elute in that order). It has been noticed that a small peak is eluting between DG and MG and this is what is starting all of my troubles.
I need to quote the resolution between DG and MG but this little peak is not letting me do this so my options are:
1: Quote the resolution between DG and the mystery peak
2: Quote the resolution between the mystery peak and MG
3: Re-process the trace so that the mystery peak is "ignored"
4: Add the resolutions obtained in choices 1 and 2.
In my opinion, if a peak is present between DG and MG then that should form part of the resolution and certainly not be ignored but it is trying to get my point across. Does anbody know where i can find clear guidelines into what to do with this situation?
Additionally, somebody else wants to change the system suitability method so that the integration parameters ignore the peak. But from my understanding, the system suit method is used to show that the system is performing ok and that the method the sample is being run on is performing well.
Sorry about the rant, i hope this is in the right section. if anybody can provide some guidance then that will be truly appreciated.
Thanks
Steve