Anthracene problems with detection
Posted: Mon Dec 02, 2019 8:15 pm
We run a mix of 36 PAH and OPAH on an Agilent 5977e with HP5 MS UI
There are varying detection limits for each compound, but the 16 PAH have always been detectable to 2 ng/ml or lower.
The anthracene peak is typically detectable down to 2 ng/ml but the area counts for ANT are at least 10x lower than phenanthrene.
But recently the ANT is now almost undetectable, we can see it at around 2 ppm, which is not acceptable. With a new column, a cleaned source, and new liner/septum/syringe there has been no improvement.
I don't think ANT should be much more difficult to detect than PHE, so what would cause such a large difference between the two?
ANT is the very small peak at 26 minutes.