Shimadzu GC MS 2010 lost communication with PC

Posted: Wed Nov 27, 2019 1:07 pm
by elenisaz
Hi everybody!
I have a problem with Shimadzu GC QP-2010 equipped with Shimadzu MS 2010 plus and controlled by a PC via software GC solution ( real time analysis).
Unfortunately after 8 years, the battery on the CPU unit of the GC died so we had to replace it with a new one. We went to GC configuration and enter all the parts involved (i.e. the injector, the MS) etc. The problem is that the PC does not recognize the GC. The GC with the MS is connected through a RS-232 cable and then a USB cable connects MS with the PC. In the "Transmission parameters" we entered
Protocol: Level 3
Baud rate: 38400 and we tried again with 9600, but nothing happened.
The PC does not see the GC unit, and the GC unit does not see the temperature of the ion source of the MS.
Is there any idea????
Thanks a lot
Eleni

Re: Shimadzu GC MS 2010 lost communication with PC

Posted: Tue Dec 03, 2019 8:23 pm
by Jake
You shouldn't have to change anything with the PC settings since the problem was with the GC. Is there a 'communications' setting on the GC that was cleared and needs to be reset?

Re: Shimadzu GC MS 2010 lost communication with PC

Posted: Wed Dec 04, 2019 8:05 pm
by antonk
GC 2010 communication options
have to be set to 115200 protocol level 3

Re: Shimadzu GC MS 2010 lost communication with PC

Posted: Thu Jul 10, 2025 1:56 pm
by ctlyohanmoreno
GC 2010 communication options
have to be set to 115200 protocol level 3
i am having exactly the same problem, but i have not found where to set that GC 2010 communications options, could you please help me?

Re: Shimadzu GC MS 2010 lost communication with PC

Posted: Thu Jul 10, 2025 2:18 pm
by ctlyohanmoreno
GC 2010 communication options
have to be set to 115200 protocol level 3
ok, sorry for my fast question about where to find the setting, i found it, but it was set already to 115200 protocol level 3, is there any other way to troubleshoot the communication?. thanks!